Hi @cherkey - Thanks for sending the issue report from the test machine. It shows exactly what is happening. There are two separate issues, plus a Joulescope UI issue that makes things much worse.
1. JS220-005556 keeps disconnecting
Over the logs you sent, JS220-005556 disconnected and reconnected about 100 times. Your other JS220s almost never did. Each time, Windows reports a USB error and the device comes back about 0.2 seconds later. The instrument keeps running during these drops, so this looks like a USB connection problem rather than a power loss. To find the cause:
Swap the JS220-005556 with JS220-001673. Only swap the instrument so that all cables connected to JS220-005556 are now connected to JS220-001673, and all cable previously connected to JS220-001673 are now connected to JS220-005556. Do the disconnects follow JS220-005556 or move to JS220-001673?
If it moves to JS220-001673, consider using a different USB port. Also share any details about the cabling for this Joulescope. Particularly ground loops, bad USB cables, or problematic hubs / hub ports.
If they always follow JS220-005556, whatever cable or port it uses, please let us know and we will look into a replacement.
2. JS220-004982 is set to 1 MHz
The configuration in your report shows that JS220-004982 is set to a 1 MHz sample rate, not 100 Hz. The other three JS220s are set to 100 Hz. Each JS220 has its own sample rate setting. To change it:
- Open the Device Control widget.
- Expand JS220-004982.
- Set the Sample rate to 100 Hz.
3. Recording issue when a device reconnects
The logs also show a Joulescope UI issue, and it is what filled your drive. When a JS220 reconnects, the Joulescope UI restores your 100 Hz setting, but only about 0.1 seconds after the device has already started streaming at full rate. During a recording, that short burst of full-rate data makes the JLS file jump ahead by billions of samples. The file then grows by tens of gigabytes with empty data, and that device records no useful data after the first reconnect. This is also the screen “jump” you saw: the waveform view resets each time the device reconnects.
We are working on a fix so that a reconnecting JS220 starts at its configured sample rate. We are also making the recorder safe against this kind of jump. We’ll post here when a release with the fix is ready.
Until then, the most reliable workaround is to fix the disconnects in step 1.
Thanks for your patience and for submitting the report!